Hi All,
i have a question regarding the vna i am using.
after calibration of the vna we will check it whether that is correct or not by looking at the smith chart.
so on looking on it after connecting a 50 ohm(same used in the calibration kit) load it shows a point at the centre.
but at the same time i was checking it's behaviour at the open condition it is supposed to be at the right corner of the horizontal line. if not then it should so some high impedance,but it is not.
without connecting anything on the cable it shows the impedance of -33+j(-2000) @1Ghz and (-7-j279) @8Ghz(calibration frequency range).
and on using the open load from the calibration kit it gives (0.1-j245)@1Ghz and (0.03+0.62)@8Ghz
is something wrong with my calibration.
i have done same thing using different cables.is it correct behavior. because for open load condition the impedance should be high. for short load condition the impedance is very low so that is understandable.am i correct to think that the impedance will be very large like in kohm for open ckt condition
- - - Updated - - -
i am using feq range of 1-8ghz for calibration.
i have a question regarding the vna i am using.
after calibration of the vna we will check it whether that is correct or not by looking at the smith chart.
so on looking on it after connecting a 50 ohm(same used in the calibration kit) load it shows a point at the centre.
but at the same time i was checking it's behaviour at the open condition it is supposed to be at the right corner of the horizontal line. if not then it should so some high impedance,but it is not.
without connecting anything on the cable it shows the impedance of -33+j(-2000) @1Ghz and (-7-j279) @8Ghz(calibration frequency range).
and on using the open load from the calibration kit it gives (0.1-j245)@1Ghz and (0.03+0.62)@8Ghz
is something wrong with my calibration.
i have done same thing using different cables.is it correct behavior. because for open load condition the impedance should be high. for short load condition the impedance is very low so that is understandable.am i correct to think that the impedance will be very large like in kohm for open ckt condition
- - - Updated - - -
i am using feq range of 1-8ghz for calibration.