My best hint is don't follow this line of investigation. A component failing after a minute with 300+ Volts across it is unlikely to show a problem with a low voltage DVM probe. You would either have to replicate the operating conditions or just replace it anyway. Given their relatively low cost I think replacing is by far the best option. Even if not 'faulty' there is a good chance they are now out of spec after all those years.I will try to search on the web, to find information about checking electrolytics but also the polypropylene HV types. If you have any hints let me know
The output voltages and the crowbar circuit should be checked first.
There are 8 different voltages involved in the crowbar triggering. 7 of them can be adjusted.
The +15.5V output can't be adjusted, and it is the voltage that is shorted by the crowbar.
The purpose of the crowbar is to blow the fuse when there is an overvoltage.
Edit: maybe R7 on the primary side should be used to adjust the +15.5V output.
SCR A2, on page 2 and page 7.What is the crowbar?
My best hint is don't follow this line of investigation. A component failing after a minute with 300+ Volts across it is unlikely to show a problem with a low voltage DVM probe. You would either have to replicate the operating conditions or just replace it anyway. Given their relatively low cost I think replacing is by far the best option. Even if not 'faulty' there is a good chance they are now out of spec after all those years.
Brian.
Measure the voltages over C22 and C23 before the problem happens. The center voltage should be stable and half of +V.
If the leakage is too high, the center voltage can slowly drift away and cause problems after a while.
My best hint is don't follow this line of investigation. A component failing after a minute with 300+ Volts across it is unlikely to show a problem with a low voltage DVM probe. You would either have to replicate the operating conditions or just replace it anyway. Given their relatively low cost I think replacing is by far the best option. Even if not 'faulty' there is a good chance they are now out of spec after all those years.
Brian.
Try to connect an old-fashioned light bulb instead of the fuse. If you are lucky the light bulb will stay off until the problem happens, and it will turn on when the fuse normally blows.
If it stays on, you can try to find the current path by measurements.
My best hint is don't follow this line of investigation. A component failing after a minute with 300+ Volts across it is unlikely to show a problem with a low voltage DVM probe. You would either have to replicate the operating conditions or just replace it anyway. Given their relatively low cost I think replacing is by far the best option. Even if not 'faulty' there is a good chance they are now out of spec after all those years.
Brian.
I would give the power supply to someone who knows how to fix power supplies. If you want to try and fix it yourself, then from what you have said there maybe no fault with the power supply, rather the adjustment of the protection thyristors maybe the problem. Try removing A2 to A5 and see if the fuse still blows.
It can be an isolation problem in the common mode choke L1. Remove it and bypass with two wires. If the scope works, get a replacement choke, or rewind it yourself on the existing core..
That would be my guess too Susan.
The fuse looks to be in the AC input to the PSU and as a fault in the output side would probably result in shutdown rather than a blown fuse, my suspicion would be the input filter up to the two big black reservoir capacitors.
It doesn't appear to have a PFC circuit so there is probably a direct connection from the big transistor's collector, through the transformer to the positive side of one of those capacitors. See if you can break the link or failing that, isolate the transistor collector connection (NOT the base or emitter unless you can do both!) or remove it completely and see if the fuse still blows. If it does, move step backwards and see if you can isolate the bridge rectifier by removing it or unsoldering both AC legs.
Brian.
update: I found this on the internet, please note the caution on it. It looks promising it is the same PSU: