T
Thanks, it is a very conductive ESD mat...if you touch it with live and neutral 1 inch apart, the mat smokes.Maybe the ESD mat is more conductive than needed and feeding the voltage from the DUT to your arm.
Thanks, the manual doesn’t say and doesn’t give isolation voltage etc.I suspect the problem is with the "isolated 230V AC". How exactly is it isolated?
Thanks, I assume this is for safety reasons?When working on high voltages, a static dissipative mat should be used, not a conductive one, especially one with such low resistance.
Thanks, that sounds a good idea though it wouldn’t be possible with this measurment as we have the grounded sheath of the coaxial probe soldered to the “non-ground” side of the sense resistor….(so that it gives a positive voltage on the scope because having negative going voltages makes things too confusing for our poor, low_IQ heads)You should use a second leakage path (like a second ESD strap wire) to the 'ground' of the LED driver so it can't float if the probe ground is removed.
…thanks, I’d like to do just that…but you cant find x10 probes that are coaxial…..well, you can get scope probes as you know, which are indeed coaxial, but they have the dangling ground clip which picks up noise……the coax x1 probe is good because you can strip the end and solder it right on to the sense resistor……yes the capacitance is not wanted but its not so bad in our particular case setup at the moment.Use a good probe instead of your 'stripped coax' or you will get unwanted capacitive effects on your circuit, ideally use a x10 probe for minimum loading.
Thanks, the voltage across the sense resistor is less than one volt so I presumed we’d be ok using the x1 coaxial probe?They are also normally rated to only 300V so be careful when using them at mains voltages
I won't worry about isolation voltage, it's surely adequate for 230 VAC. But what's the leakage current respectively insulation resistance specification?Thanks, the manual doesn’t say and doesn’t give isolation voltage etc.
We know its isolated because I connected earth to live_out via a 1 amp fuse and turned the output up to 230vac and the fuse didn’t blow..therefore we know its isolated.
Thanks, the manual of the AC PSU didnt say.I won't worry about isolation voltage, it's surely adequate for 230 VAC. But what's the leakage current respectively insulation resistance specification?
Thanks, but i must add that the coaxial probe, which was connected to the scope via its BNC connector, was , at its other end, soldered to the PCB at the sense resistor, so therefore the PCB of the LED driver should have been at earth potential...at least at the point where the earthed sheath of the coax probe was soldered to it.Since it was isolated, the power supply ground connected to the DUT was at a potential higher than your earth ground
Double check this - based on your earlier posts you make a conflicting statement.The sense resistor is in the negative_DC_bus, just downstream of the mains rectifier.
As you know, the BNC of the coaxial probe is connected to the scope, so the sense resistor “should” have been at earth potential.
treez said:Thanks, I assume this is for safety reasons?When working on high voltages, a static dissipative mat should be used, not a conductive one, especially one with such low resistance.
Thanks but sorry i do not understand what is meant here. To clarify, i do not know if neutral_out of the isolated AC PSU is connected to earth potential. I will check this. I do know that live_out of the isolated AC PSU is not connected to earth....oh sorry,... i actually did a test whereby i connected the live_out to earth (via a fuse) and switched the isolated output on at 230VAC......so this confirms that both live_out and neutral_out of the isolated AC PSU are isolated from earth ground. As such, i cannot be shorting out any of the diodes in the mains rectifier.You state that one side of the isolated AC is grounded (that's fine)
Thanks, though i wonder why the conductive ones we have are for sale?You should use a static disipative mat as a conductive one will still transfer charge quickly, which is exactly what you don't want for any kind of ESD protection.
After shocking myself again with the setup of the top post, i am beginning to wonder if the scope socket is actually at earth porential after all....maybe the earth wiring in our building does not actually go to "planet earth" (eg via an earthing rod), and maybe the "earth" connection is floating at who knows what voltage?
....maybe the earth wiring in our building does not actually go to "planet earth" (eg via an earthing rod), and maybe the "earth" connection is floating at who knows what voltage?
thanks, it has a 3 pin plugDoes your scope and AC PSU have a 2 or 3 pin mains plug on them?
Thanks, its a Lisun Group AC PSU, but they dont sell it any more, and there are no details on it..its not one of the following, and must have gone obsoleteI won't worry about isolation voltage, it's surely adequate for 230 VAC. But what's the leakage current respectively insulation resistance specification?
..Thanks i appreciate that if its floating, but in the case of the top post, it wasnt exactly floating any more because i earthed it at a point with the scope due to the coaxial scope probe connection.You need a leakage current below 1 mA to safely touch the floating circuit.