Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Why use notch filter but not a attenuator?

Status
Not open for further replies.

rhjang

Junior Member level 3
Junior Member level 3
Joined
Nov 4, 2003
Messages
29
Helped
2
Reputation
4
Reaction score
0
Trophy points
1,281
Activity points
339
When measuring GSM Tx spurious, the reason of use notch filter is that signal analyzer's dynamic rang is not enouugh.

I use R&S FSIQ7 which have noise floor at -88dBm(RBW 3M), and I want to measure a 30dBm EGSM signal.

First I use a notch filter to filte out main signal and measure the 2nd harmonic, all lose are considered.The 2nd harmonic is about -40dBm.

Since the 2nd harmonic is -40dBm and the GSM signal is 30dBm, if I use a 20dB attenuator ,the main signal and 2nd harmonic into the analyzer are 10dBm and -60dBm, these signal are in the analyzer's dynamic range.
So I use 20dBm attenuator, level offset set to 20dBm to shift back the signal, and measure in the same RBW, but the 2nd harmonic is greater than -20dBm!!

Why the measured 2nd harmonic is so different :!: :?:
 

This usually happens because the 1st mixer is saturated by the carrier signal ( even if it is not existing on the screen, the 1st mixer is very wide band device ) and the mixer itself generates the harmonics.
Try this : increase the imput attenuator by 10 dB and check if the harmonics level decrease: if it is so the mixer itself is generating harmonics.