I would like to determine conductivity of water with complex impedance measurement based on AD5933. I do not have very good electro-chemical background so I am not sure how to connect impedance (magnitude, phase angle) with conductivity.
The dissipation factor or Loss Tangent or complex impedance of water can be done with high voltage if in the GΩ range, but constant current is best method then use PLL quadrature detector to measure VI directly or use phase angle vectorfrom a phase detector and measure amplitude of voltage from CC source.
. Omega Meters tend to use high voltage for board range of leakage resistance which is important due to ionic and particulate matter contamination such as sodium and metallic particles. The dielectric constant is fairly constant but varies with frequency. Instruments have been designed to measure complex impedance from 0.01 Hz to 1MHz to measure the water content in mineral oil used in HV transformers, but 10 Hz gives useful results or far away from AC stray noise. (AC Notch filter or tuneable BPF and very high CMMR may be required. The conductance per unit volume is the std method I believe.
Maybe you tell us a little bit more about your application, you want to measure the cond. of water in a can or of the ocean ? Should it be a continous measurement or just at specific time periods (e.g. one time per day). Do you need a very precise measurement?
If a DC voltage is applied, ions would build up an "ion cloud" near the surface of the electrode. This would prevent other ions in solution from being attracted to the electrode.
This is referred to as the "polarization effect". Therefore, AC potentials are utilized to prevent localized build up of ions. The AC driven signal can be operated between 2-10 volts at 60 Hz-3,000 Hz, with lower frequencies being used for lower concentrations like 10Hz.
I have attached my results which I obtained today.
I used simple structure of two copper lines separated on FR-4 substrate to have constant distance between electrodes. Using my complex impedance system I varied frequency in range from 5 to 100 kHz (constant temperature).
This is magnitude change
and phase angle change
How I can measure conductivity based on this results? For example, I will use distilled water or other liquid with known characteristics for calibration but how later to determine on unknown liquid?
I am not able to perform sweep beyond 1 kHz because I am using self-designed impedance analyzer based on AD5933. Plot when capacitor is in air is very huge resistance, in range of MOhms and GOhms... I am not sure if I can calculate capacitance of structure with simple equation like I presented in post
Yes, I will make temperature compensation of obtained results. But I still do not have idea how to connect complex impedance with permittivity/conductivity?
Did you check correct operation of your impedance measurement by connecting a fixed resistor or RC parallel circuit?
Presuming the results are correct, it's obvious that they can't be converted to frequency independent complex impedance or permittivity. You can of course calculate the elements of an equivalent RC circuit for a given frequency. The phase deviation is however small enough to take the impedance magnitude numbers as resistance and convert it into conductivity, preferably for a frequency in the standard low kHz range.
I don't know the geometry factor of your conductivity cell and what the 2 kOhm numbers mean in µS/cm terms.
Yes, I checked operation of complex impedance measurement system with a lot of different RLC combinations. Magnitude error is less 1 % while phase angle error is about 1.5%.
I would like to convert it to frequency and temperature dependent complex permittivity.
So you are just asking about the arithmetic? It should be obvious from definition of complex permittivity, except for the cell geometry that must be empirically determined.
I presume you have no problems to convert the impedance measurement from polar (magnitude, angle) to cartesian coordinates (Z = R + jX) or admittance (Y = 1/Z = G + jB).
The cell geometry I will determined with distilled water, for example.
But how to move on? It is not a problem to move in Y = 1/Z = G + jB from Z and Theta, but still I can not reach how to connect conductance and susceptance [Siemens] with complex permittivity?
This will helps a lot but I still can catch how to estimate A, even if I am using a liquid with known dielectric permittivity and know dimensions of capacitor.
I will calculate G and B (based on R and X) and then I have
σ=A/G (A is still unknown)
ε'=-B/(Aω)
I can not estimate A from the first equation because I do not know σ...
You claimed to calibrate the cell with distilled water...
In any case you either need to know the geometry factor (e.g. from a numerical field calculation) or calibrate with a conductivity standard. KCl solutions with e.g. 100 µS/cm or 1000 µS/cm are commonly used. Distilled water would be too inaccurate I presume.