The load is 12V 500W 41.67A heater.
yes, for safety we are running on 12v dc, can't use transformer, as it is adding cost and cause magnetic problems, we are using in medical device'Heater' implies resistive element. Usually this can run on AC.
Perhaps you can get by with a plain transformer instead of SMPS?
Or does it absolutely need DC?
SMPS is not producing over temperature, nor voltage breakdown in voltage logger.The question is: Why does it restart?
Overtemperarture?
Mains voltage breakdown?
.. I mean: under normal conditions (see SMPS datasheet) it never should restart.
Also: check whether the SMPS is able to start under full load?
Klaus
Makes no sense to me.The SMPS supplier have made the point the low resistance may be the reason.
Either:Makes no sense to me.
If it is specified for 12V / 83.3A then it needs be able to handle load resistance down to 144 mOhms.
I´d ask them how the SMPS can be specified to deliver 83A while it has problems even with 36A.
It´s just Ohm´s law.
Klaus
They are asking to add series diode to increase resistance,Either:
1) The person you talked to at the mfr doesn’t know what they‘re talking about. Or:
2) The mfr lied about their load capacity.
Maybe you could also try tying a banana to the power cable.They are asking to add series diode to increase resistance,
I agree with the other posters, especially barry here. So long as you stay within the rated specifications of the PSU, the resistance of the load is irrelevant. The suggestion of adding a series diode in particular is silly.Either:
1) The person you talked to at the mfr doesn’t know what they‘re talking about. Or:
2) The mfr lied about their load capacity.
As an aside, this does not make sense, at least according to medical device safety standards I'm familiar with. Is your PSU isolated? If so it has a transformer in it. Not sure what you mean by "magnetic problems".yes, for safety we are running on 12v dc, can't use transformer, as it is adding cost and cause magnetic problems, we are using in medical device
This has absolutely nothing to do with what the OP posted. OP has already stated that it runs fine for 15 minutes and THEN fails.Hi
Can you make clear your test case ?
I guess you connected 0.33R to power module then turn on AC input.
If you do like that, the power started with heavy load. Each SMPS power module limited peak output current or output capacitor or soft-start time, if these condition are not met, controller will consider has something fault like shorted-circuit and try to start-up. Because, when SMPS start, it need to charge output capacitor and 0.33R load, so the peak output power that time is higher than the limitation, module will reset/restart.
Can you try to start SMPS before, then plug load 0.33R.
If module can run normally, it mean it is still OK. Maybe, the design don't use current mode control.
If module can't run, it maybe faked its output power. You can try to use higher load as 1R, then reduce step by step to see what real output power is. Or its current sense circuit has some problem.
So now in total 3 SMPS?Tested with two brand SMPS, the second one, branded costly one restarts in a moment,
but the cheaper one switch off after 15min.
Without additional information .. this sounds like nonsense.The heater manufacturer also suggests transformer PSU, but they told they have inhouse R&D which can develop an SMPS to suit this particular heater.
What are "this kind of heaters"?they also telling this kind of heaters behaves like this