The schematic is unclear in several regards. How do you connect thermocouples to the differential input, how's a suitable common mode bias achieved?
Thanks FvM.
The circuit has 2 analog inputs in which we can connect RTD, TC and 4-20mA.
J3 connections:
1. Analog input 1+
2. Analog input 1-
3. Analog input 2+
4. Analog input 2-
5. Analog inputs common connection (used for RTD measurements).
6. System GND (used for 4-20mA measurements)
I have attached drawings for connections based on sensor type.
For TC measurement, U13 AD7792 generates and connects the bias voltage (AVDD / 2) on AIN1(-) or AIN2(-). U13 has internal reference.
For RTD measurement, equal excitation currents are generated by U13 on IOUT1 and IOUT2. These currents are fed to Channel 1 or 2 through U14, which has 2 x SPDT switches. The equal currents cancel wiring resistance on the measurement and return through J3-5, where they produce the reference voltage on R9 for RTD measurements. The 100nF was added because the same noise problem existed also on RTD measurements but canceled by filtering reference voltage also with this capacitor.
SJ1 on U13 is just a jumper made up of pads on the pcb to account for DG2731 and DG2733 differences, so assume that 2-3 terminals are shorted.
For 4-20mA measurement, S4 and/or S5 shorts 2-3 and 5-4 terminals so the current is converted to voltage. Voltage is added through 33 Ohm resistor on the (-) side of the input because U13 needs any absolute input voltage to be at least 100mV above GND for correct offset and gain results.
All resistors on analog paths AIN1(+/-) and AIN2(+/-) and U14 are used for current limiting except the ones together with capacitors just before U13 inputs, which are used as low pass RC filters on signal.
Some more facts about the circuit:
- Regarding any possible ground problems. Internal plane 1, is the GND plane. It is separated from the GND to the rest of the board and connected only on the upper left corner as shown in the picture. Although I have set the minimum clearance from GND plane to 0.7mm, I believe that the resulting area is sufficient for the analog node which draws about 5mA maximum during conversion. Please correct me if I am wrong. But I am not sure if I have joined analog node GND plane to the rest of the PCB GND plane correctly. The connection path remains on the internal plane and it is 5mm wide.
- I noticed that if I switch on a CCFL light, the circuit picks up noise during light startup.
- By removing U14 (analog switches) the problem is canceled and thermocouple measurement becomes also very stable on CCFL light EMI.
We can assume that U14 injects all the noise to the signal path. But then it is connected to the same VCC plane as U13. Is it possible to inject so much noice from power supply on the signal path while U13 is not ?
We can also assume that the noise on VCC exists on all signal paths and is cancelled at U13 because of the differential design of the measurement. But then why analog switch inserts noise ? Does this has to do with phase change issue or any mismatch between the two indepentent analog switches that creates differential errors ?
Dimitris
- - - Updated - - -
All resistors on analog paths AIN1(+/-) and AIN2(+/-) and U14 are used for current limiting except the ones together with capacitors just before U13 inputs, which are used as low pass RC filters on signal.
Dimitris
I mean current limiting for overvoltage protection.