Hi,

I want to know that for load pull measurement (on wafer devices), do we get the power contour at constant (or 1dB compression point) delivered input power or constant absorbed input power.

i always find that paper only use Pin without specification.


thanks in advance.
 

Re: load pull measurement

Its the delivered power.
Mazz
 

Re: load pull measurement

The countures are the output power for let say 1dB steps. The input is not matched to 50 ohms and the input power sometimes might be high because the input R.L is bad. So it is for constant input deliver power but the issue is the ouput and not the input power.
When you simulate load pull you can match the input port for maximum absorbing power or you can just increas the input power in order to get the output power you look for.

D.J
 

Re: load pull measurement

The problem I meet is that the input R.L is bad, just as D.J mentioned, and due to the limitation of our measurement system i can not do source matching, can I still use fixed dilevered power to find the optimial load, with the absorbed input power changes much at different load?

another thing is how to choose the input power level for finding optimal load?
It seems that for the focus load pull, it chooses the Pin at 1db compression point for very load.
Is there any system choose fixed Pin for every load?

thanks

D. X
 

Re: load pull measurement

Look for papers by Andrea Ferrero, I think he is the major expert on load pull (both active and passive setup ) in the world.