Okay, I'll try to add one notch filter of 50Hz.In my post about filtering there is a thought that environment is high common mode (CM)
so might be best to place notch filter first so that you do not G up the CM and possibly
saturate the filter signal path. Also take G in it for noise cascade reasons.
Okay, I'll try to add one notch filter of 50Hz.In my post about filtering there is a thought that environment is high common mode (CM)
so might be best to place notch filter first so that you do not G up the CM and possibly
saturate the filter signal path. Also take G in it for noise cascade reasons.
Thank you for the correction, I have changed the capacitor value to 0.5nF to both C2 and C6. I have also put a Capacitor near the battery supply in the real circuit. And also I have not put 100nF near the microphone input, that was only for the simulation. And 2 batteries here are just in the simulation only not in the real circuit.The article says there are no breathing sounds above 2500Hz so maybe your cutoff frequency should be 2.5kHz. It is only 100Hz in your schematic.
I cropped your schematic to it is not as big as my neighborhood and made some corrections and notes.
Some terminology issues.I have one more issue left to resolve now, to give the signal to ADC, which I think I need to shift the signal a bit higher in order to capture in ADC. And plotting issue is still there because the breathing pattern is hardly received from other chest portions, so the comparison is a little time taking as we don't have patients to test.