Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

output changes when I touch the enclosure

Status
Not open for further replies.

hemnath

Advanced Member level 3
Advanced Member level 3
Joined
Jun 24, 2012
Messages
702
Helped
61
Reputation
120
Reaction score
57
Trophy points
1,308
Location
Chennai
Activity points
6,589
Hi,

I have a transmitter board which senses the pressure and outputs 4 to 20mA. The board is mounted in a aluminium enclosure using the Stainless steel standoff. The screws used are Stainless steel.

Normally, the board is working good and provides 4 to 20mA based on the input pressure. But when I touch the aluminium enclosure, the output current value changes to 3.96 mA from 4mA.

Why is this behaving strange?
 

i see two possibilities

by touching the enclosure, you have somehow reduced the pressure and so the current goes down
0.04 mA is likely well within the design margins of the pressure transducer and the output 4 to 20 mA

by touching the enclosure, you change the local distributed parasitic impedance

it is quite possible there is another explanation

please provide a schematic and mechanical layout
 

I would guess that in an industrial environment, with an
ungrounded or poorly grounded (i.e. "earthed") case, you
are adding 50/60Hz hum to the local "ground" and some
rectification / charge pumping (like with input ESD diode
networks) is adding to the real signal.

You might repeat the experiment with the case tied hard
to a known good earth ground, and in a location without
much electrical equipment, fluorescent lighting and so on.
You could put an oscilloscope to the internal input point
and eyeball the waveform as you touch / un-touch the
case, to get quantitative. If you've ever just touched the probe
tip on a 'scope you have an idea of how large human body
line frequency "pickup" can be against a high impedance
sense-point.
 