Continue to Site

Welcome to EDAboard.com

Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Oscilloscope shows extreme variation in frequency

Status
Not open for further replies.

raiya_23

Member level 4
Member level 4
Joined
Oct 31, 2009
Messages
73
Helped
3
Reputation
6
Reaction score
2
Trophy points
1,288
Activity points
1,717
Hi I am using a Hameg HMO3522 Oscilloscope. I wanted it to display a 350Mhz RF which I was giving through my Rhodes and Schwarz signal generator but the scope is showing extreme variations in frequency from 9kHz to several Mhz, even the voltage is not being displayed correctly although I performed the probe adjustments etc.

What could be the problem?
 

The problem could be aliasing, if you CRO can not get samples of the input waveform fast enough. One thing you could try is to switch the CRO to "envelope" - if this mode exists, it should then display a bright filled in rectangle, from which you can measure the p-p voltage of the waveform, if this is varying you have a loose lead?
Frank
 

The Hameg HMO3522 is a 350MHz oscillator. By using a 350MHz signal, you're using it at the very limit of its specifications.
Use a 200MHz-300MHz signal and see if that is displayed properly.
 

In the manual it is mentioned that you can measure signals upto 400Mhz if automatic triggering is used
 

Status
Not open for further replies.

Part and Inventory Search

Welcome to EDABoard.com

Sponsor

Back
Top