I think we are avoiding (evading?) the original question:
NO, do not use the ratio cursors if the scope has them. I assume you mean % markers on the graticule, if so they are used for something else altogether.
The way to do it to use one probe, preferably a 'x10' type which has been properly compensated. Measure the amplitude of the waveform using one of these methods:
1. Make sure you know where the zero line is by switching the Y input to ground and adjusting the Y position to overlay one of the horizontal lines on the graticule.
2. Switch the Y input to DC coupling. Now measure the voltage as the number of squares the line deflected multiplied by the V/division on the Y amplitude control.
3. Do steps 1 and 2 for the input of the summing network and the output of it. Use a calculator to work out "measured output / measured input", this gives the gain.
or if you want to use the scope 'on screen' cursors:
1. set the scope to show the difference between cursor voltages.
2. set one cursor on the bottom of the input signal, the other to the top of the signal, if you are measuring DC, set the bottom cursor to show 0V. Note the top cursor reading.
3. repeat step 2 on the output signal, again note the top cursor reading.
4. Use a calculator to work out "measured output / measured input", this gives the gain.
If your summing amplifiers have steady DC inputs, it is easier and probably more accurate to use a DVM to measure the voltages. In any case, if you are fault finding there probably isn't any need to check gain at all. If the resistor values measure OK the stage is almost certainly working properly.
Brian.