A standard 10:1 passive probe has an input capacitance of 10 to 15 pF, resulting in a 200 to 300 MHz low-pass corner with a pure resistive 50 ohm source. So "won't significantly disturb" isn't right if you want e.g. 500 MHz bandwidth. The real problem is however probe connection inductance. Any connection method where the probe tip isn't coaxially plugged to a 50 ohm connector involves an considerable amount of inductance and ruins the signal quality. Who ever tried to monitor a fast edge digital signal with a passive probe knows about the problem.The 10:1 probe should have a high enough impedance that it won't significantly disturb the signal from a 50 ohm source. But you can't have any significant length (more than a couple inches) of unterminated coax connected to the source or you will get reflections.
Yes, I made an incorrect assumption about the probe impedance at 500MHz.A standard 10:1 passive probe has an input capacitance of 10 to 15 pF, resulting in a 200 to 300 MHz low-pass corner with a pure resistive 50 ohm source. So "won't significantly disturb" isn't right if you want e.g. 500 MHz bandwidth. ...............................
Common mode rejection is another thing. The suggested dual probe method works fairly for low and medium speed signals, but has no satisfying results in the 100 MHz range.
.........................................
For low source impedance, you can also use a 10:1 or 20:1 resistive probe, a 450 respectively 950 ohm series resistor at the end of a 50 ohm cable. It can achieve several GHz bandwidth and costs about nothing when soldered in place.
I presume, DC common mode won't require differential measurement at all, just a trace offset or AC mode. I was expecting high frequent CM signals, but the OP didn't tell exactly.I can see common-mode rejection is important for a signal for with common-mode noise, but will that be a problem for the OP's signal with a constant 1V common-mode voltage?
I suppose it is obvious but it should be emphasized that when using the 50Ω cable and the 10X or 20X (-50Ω) series resistance, the cable must be terminated in 50Ω at the scope input. It should also be noted that this will result in a 10% (with 10X) or 5% (with 20X) amplitude loss from a 50Ω source.
Resistive probes require a 50 ohm termination at the oscillsocope, it's a built-in standard option for oscilloscopes > 300 MHz anyway. The signal attenuation is of course the same as with a high impedance passive probe with respective divider factor.I suppose it is obvious but it should be emphasized that when using the 50Ω cable and the 10X or 20X (-50Ω) series resistance, the cable must be terminated in 50Ω at the scope input. It should also be noted that this will result in a 10% (with 10X) or 5% (with 20X) amplitude loss from a 50Ω source.
As previousl mentioned, how should we know if you can design it? It's at least worth a try.Do you think i can design a simple Instrumentation amplifier by using 3 wide band op amps for e.g. from TI?