In scope probes, R1C1=R2C2 but since the R ratio is unknown unless you have done a leakage test to verify it is much more than 1 GOhm . Is it possible your insulator leakage is 700MHOhm?
Ideally the R Ratio would match the Cap ratio for broad spectrum and low frequency , otherwise R ratio will over-ride the cap impedance ratio at this low frequency.
The second problem is your 4.7pF Output is being loaded by the measurement capacitance of the probe and amp.
What is the frequency of the signal?
If you are not using a 10:1 scope probe and the frequency is near line frequency, then the typical scope input impedance of 1 megohm is loading down the capacitive voltage divider. You should use a 10:1 probe for the measurement. The scope input capacitance is not a problem since it is in parallel with the C2 capacitor and is much smaller than 4.7nF, however the connecting cable capacitance could be a factor if you have a long coax cable.
It also appears that the loading of the oscilloscope is loading the circuit under test which causes a frequency change. Again you should be using a 10:1 probe, if you are not.
With a circuit like this, do not hang anything across the whole tuned circuit, because you will damp the Q and detune it with the added capacitance. Just measure the 4.7 PF and 4.7 NF carefully to get their ratios. Scopes are not precision voltage measuring devices use a valve voltmeter (or its solid state equivalent).
Years ago I used this system to calibrate real valve voltmeters against a lab standard, by using a very high Q coil (Q > 470), a 100 PF and a 10 NF capacitors and a standard lab RF signal generator. I could get > 500V @ 1MHZ across the coil.
Frank
With a circuit like this, do not hang anything across the whole tuned circuit, because you will damp the Q and detune it with the added capacitance. Just measure the 4.7 PF and 4.7 NF carefully to get their ratios. Scopes are not precision voltage measuring devices use a valve voltmeter (or its solid state equivalent).
Years ago I used this system to calibrate real valve voltmeters against a lab standard, by using a very high Q coil (Q > 470), a 100 PF and a 10 NF capacitors and a standard lab RF signal generator. I could get > 500V @ 1MHZ across the coil.
Frank
We don't know anything about your objective, nor input or output impedances.
All we know is you are tuning 4.7pF with 0.5 & 1MHz
, somehow, but we don't know how or why.
Obviously anything in the pF range is easily loaded but at 1MHz, but you are looking for a 30% measurement error.
for reference @1MHz
1pF = 160 kΩ (= 10e6/6.28)
and 1000pF = 160 Ω
Thus the cap ratio dominates any leakage ratio or ESR which can be neglected at 1MHz ( but not 60Hz, which I assumed at first)
However there may be other capacitance on your layout not shown in your schematic which can introduce errors.
In order to design a fixed cap divider ratio, all leakage capacitance must be known, or you use a known signal level and determine where the stray leakage is coming from then determine your equivalent capacitance and whatever ratio... e.g. 700:1, then only use the monitor.
I do not know what your "matching" network is mean't to do. First of all short out C4 and remove R1, they only waste power and provide no useful purpose. Given the values from your circuit, the input section is 42 PF + 530µH and the output section, 70 PF +268µH. These sections both resonate at about 1 MHZ, so the input section looks like a short circuit !! So the output impedance of the generator is across the main circuit. I think to get the best bang for your buck would be to put a 70 X 50 = 3500 PF capacitor in series with the earthy end of the 70 PF and feed your RF in here. You should get loads of RF across the main coil - Like I did see above.
Frank
There are some very critical values to resonate exactly at 750kHz and 1500 kHz. The cap divider ratio is not critical as your middle cap C3 = 70pF dwarfs the 4.7pF. Note I increased the value to tune it.
If YOU MUST verify the cap ratio, DO it with standard 50 Ohm generator and 2 caps only, otherwise just use predicted ratio.
Then tune the other parts to optimize the center freq for each. Notice that I put a grid resistor value of 100 Ohm for the source. the lower the higher the Q> anything higher than 500R combines the two resonances into one in between. What do you use?
Note L ratio should be 2 and note the other critical C values changed.
Total peak gain with 100R source is >25dB balanced for each. How much gain do you need?
I have been giving your problem some more thought. Your output voltage is dependent on the Q of the output circuit,which is primarily determined by the loss of the inductor and any resistive loading. I think the reactance of the 268µH at 1 MHZ is 1800 ohms, so if its got a Q of 100, there is a loss resistance of 1800 X 100 across the tuned circuit = 180 KΩ, in addition you have a 1M5 across it. The load is just capacitive, lets say the impedance of the circuit is 180KΩ -7% ~ 165K. So you have to match your 50 ohm PA to a load of 165K, or ~3000 :1 so the tapping point should be 1/√3000 or 1/55, as your tuning cap is 70 the bottom cap should be 55 X 70 = 3850 PF. My guess at 3500 PF seems pretty accurate. Of course if your Q of the inductor was 200, the loss resistance would be twice as much, say 360 K, less the 1M5 = 280 K, the Z match, 280K/50 = 5600, tapping = √5600 ~75, so the capacitor would now be 70 X 75 ~ 5000PF . . You can see where this is going
Frank
6kV into 1.5M load is presumably non-inductive carbon or ceramic 25W load?
How did you make a zero Ohm source impedance?
If you slide the cap values you can see the spectrum shift almost instantly. on my simulator.