Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

confusing about Rogowski coil measuring & calculating

Status
Not open for further replies.

daffdaff

Junior Member level 1
Junior Member level 1
Joined
Dec 23, 2015
Messages
19
Helped
0
Reputation
0
Reaction score
0
Trophy points
1
Activity points
173
Hi Everyone,

I've made the Rogowski coil for current pulse dectection, and it's able to work.

I have two questions.

1. How do I know the scale factor of my coil (mV/A) ?

2. If I use numerical integration (trapizoidal tool in Matlab) to integrate the output signal of the coil (data point from oscilloscope). The result is primary current already right?
( Vo=M(dI/dt) ---> I=(1/M)∫Idt ) or I still need to use scale factor of my coil (mV/A) ?

Thanks for answer.
 

The simple equation for Rogowski coil sensitivity can be found everywhere in literature, e.g. here https://en.wikipedia.org/wiki/Rogowski_coil

Rogowski coils is a dI/dt sensor, the scale factor must have a unit of V*s/A not V/A.

Multiplying the scale factor with the integrator "gain" 1/T gives the overall sensitivity.
 

Yes, the scale factor that you told is M (Vs/A) right? (multiple with dI/dt, we will get unit in Voltage). But, mine is not.
Rogowski coil is current transducer but it give the output in unit of voltage. So the scale factor that I mean is some factor that multiple with output of the coil(Voltage) = Current that it was measured (A)
 

From this link, You will see output sensitivity in mV/A
**broken link removed**
 

Untitled.png

It is the commercial Rogowski coil fluike i6000s, and it has scale factor in unit of mV/A. It make me confuse with M (Vs/A)
 

The shown Rogowski coil sensor has a built-in integrator, thus it has mv/A scale factor (M/T).
 

Ohhh I get it. Thanks a lot, and I'm sorry to waste your time.

1. If I use numerical integration, the scale factor is only M.
2. If I use an external integrator, the scale factor is (M/T).

Is that correct?

For the second choice, How do I know the value of L or C of integrator, if I want the result to closely match shape with measured pulse.
 

The Rogowski sensor with built-in integrator basically derives the original current waveform. The instrument has upper and lower bandwidth limits given in the specification. The exact low frequency cut-off characteristic and related waveform deviations have to be determined empirically.
 