Chop mode is design to lock to not sync to the incoming repetitive signal but lock to an offset with at least 100 Hz difference frequency to create a smooth trace.
Alt mode synchronizes to the selected polarity from the selected source and with retrace blanking , may or may not be synchronous to the incoming signal. THis problem of an injection locked loop or pleiochronous syncing causes potential correct image capturing of all channels.
I'm sorry, I really don't understand what you are trying to say there.
A scope's operation is really quite simple...
There is a trigger, which is derived from channel A or channel B or ext trig. This is completely independent of chop/alt mode.
When the trigger occurs in chop mode, both traces are alternately "sampled" at the chop frequency, and displayed one above the other during the same sweep. Yes, that is a perversion of the word "sampling", but it serves to highlight that time differences less than the sampling/chopping frequency are invisible (only one channel is displayed at any given time).
When the trigger first occurs in alt mode, channel A is displayed for the complete sweep. On the next occurrence of the trigger, channel B is displayed for the complete sweep. Provided that the signals on channel A and B are repetitive, the signals on channel A and B therefore shown in the correct relative positions on the display. That means that any time differences within the scope's front-end analogue bandwidth will be visible.
Note that alt mode is even useful when the signals are not repetitive. Providing the trigger is stable, the "
eye diagram" is displayed. Note also that the eye diagram is a key specification
and test method for high-speed signals, where signal integrity need to be assured.
- - - Updated - - -
I believe that the A+B mode is the best.
That entirely depends on the signals being measured. A universal statement has to be wrong.
Like wise the Alt mode leaves leaves you with the uncertainty of the trace flyback time.
No. You misunderstand how scopes work.