Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

A problem with a used scope probe compensation

Status
Not open for further replies.

OutputLogic

Junior Member level 2
Junior Member level 2
Joined
Jun 11, 2009
Messages
23
Helped
1
Reputation
2
Reaction score
1
Trophy points
1,283
Location
San Jose, CA
Activity points
1,370
I've purchased a used P6114B probe for TEK TDS-380 oscilloscope.

I cannot properly compensate it - there are significant overshoots on a square wave compensation signal.
Other probes I have work fine with that scope.

Here is a screenshot : ch1 is a bad probe compensation signal, ch2 a good one


Can anyone suggest what might be the problem?
 

Your report seems to indicate a defective probe, doesn't it?
 

I have some marginal ones that act this way, check the ground
at the 'scope input BNC (if I wiggle it there, it gets better on
one of them).

That trace looks like you have excessive DC attenuation,
which a poor ohmic connection would produce.
 