Welcome to our site! EDAboard.com is an international Electronics Discussion Forum focused on EDA software, circuits, schematics, books, theory, papers, asic, pld, 8051, DSP, Network, RF, Analog Design, PCB, Service Manuals... and a whole lot more! To participate you need to register. Registration is free. Click here to register now.

Markers for measuring frequency and amplitude in ols spectrum analyzers and scopes.

Status
Not open for further replies.

neazoi

Advanced Member level 6
Advanced Member level 6
Joined
Jan 5, 2008
Messages
4,157
Helped
13
Reputation
26
Reaction score
15
Trophy points
1,318
Location
Greece
Activity points
37,198
There is a problem with old analogue scopes and analogue spectrum analyzers and this is the absence of markers. I face the problem with my tek 491 and this is the reason I do not use it, because I cannot measure signal levels and frequencies reliably.

I want to use this old gear for basic amateur work, which can really go to 40HGZ with it's plugins.
I was thinking that a marker can be done by having a relay and altering between the signal under measurement and a calibrated signal source. When the relay is in the signal position, it will feed signal to the SA. When it is in the source position, it will feed the generator signal to the SA. By alternatively switching these two and varying the frequency and the amplitude of the calibrated signal source, one could find frequency of the measured signal and tit's amplitude with a good approximation, much better than "counting squares".

In the time domain (oscilloscope) markers are not that easy. But again, one can alter between a calibrated signal source and the under-measurement-signal and compare the screen amplitudes. That way you can have a good aproximation of the actual peak to peak amplitude. in different parts of the measured waveform, even if it is not a pure sine.

What do you think of my idea? Has anyone tried to do the same thing before? Are there any similar projects out there?
 

The theory is OK but the practicality of using a relay makes that idea unusable. Assuming your frequency sweep is fast enough to appear continuous, the relay would have to switch at twice that speed. You are probably looking at switching at least 20 times per second and the switchover has to be synchronized to the sweep. The same speed is needed for amplitude switching, essentially you are swapping between the normal display and a reference display at sweep frequency.

You could possibly do it with a semiconductor switch but then you would have to be careful with bandwidth and voltage capabilities, especially at GHz frequencies.

If you can get access to the intensity control, a better solution would be to brighten up a spot on the trace and from the horizontal position, calculate the frequency and simultaneously "sample and hold" a snapshot of the amplitude. It requires some circuitry but the sweep waveform is already available and measuring a voltage will be more accurate than visually aligning two alternating sweeps.

Brian.
 

    neazoi

    Points: 2
    Helpful Answer Positive Rating
The theory is OK but the practicality of using a relay makes that idea unusable. Assuming your frequency sweep is fast enough to appear continuous, the relay would have to switch at twice that speed. You are probably looking at switching at least 20 times per second and the switchover has to be synchronized to the sweep. The same speed is needed for amplitude switching, essentially you are swapping between the normal display and a reference display at sweep frequency.

You could possibly do it with a semiconductor switch but then you would have to be careful with bandwidth and voltage capabilities, especially at GHz frequencies.

If you can get access to the intensity control, a better solution would be to brighten up a spot on the trace and from the horizontal position, calculate the frequency and simultaneously "sample and hold" a snapshot of the amplitude. It requires some circuitry but the sweep waveform is already available and measuring a voltage will be more accurate than visually aligning two alternating sweeps.

Brian.
That last paragraph, reminds me of an old Tek SA where they introduced a notch at a movable point on the screen, then measured the frequency of the notch with a mehanical meter.

Why do I have to switch the relay that fast?
Actually, there will be no automatic switching at all, but a user switch (eg with a button). The user will switch the two waveforms up to the point where their amplitudes/frequencies are the same. The eye can easily track the positions of the waveforms that appeared on the screen a few ms ago and have now vanished and replaced by the other waveform.

In the Tek 491, which has a P7 phosphor with high persistance, this is even easier. In fact these oldies were based on persistance as a form of analogue "video memory", so as to operate the sweep oscillator at a lower rate and achieve higher resolution, yet draw a usable line onto the screen.

But the same idea can be usable in screens without long persistence because of the eye/brain persistence of vision. It is actually easier than it sounds. For example when you see two dots of alternatively switched laser pointers, you can easily bring them at the same position.

With this idea, you avoid all the hassle of interfering with the internal circuits of the scopes/SAs and it is more universal, although not really accurate like you said, but yet quite usable, far more than counting squares.
 

this is one of the reasons people used waveguide frequency meters, back in the day. If you were not sure what frequency something displayed on the screen was, you tuned the frequency meter to it, noticed when the amplitude dipped, then read the frequency off of the meter's markings
 