Hi everyone
I am designing a product which needs an isolated measurement input which will connect to mains (as well as various CATII circuits up to 1kVrms).
I am designing to IEC61010 and need to get communication across an isolation barrier as well as power across the isolation barrier to the analogue measurement circuits (as well as some digital circuitry).
IEC61010 describes the overall product testing requirements and creepage/clearance etc but it does not detail the requirements of the isolators themselves.
There is are a number of standards that apply to the opto isolators (I will use opto isolators to transfer data across the barrier), such as the IEC Standard is IEC60747-5-2.
The Broadcomm ACNV4506 opto is good for an operating voltage of 2262V, I need 2kV isolation so this is perfect.
However, I am not sure which standard to refer to for the isolated power.
I think I will need to design my own isolated PSU as i can’t find an off the shelf model offering a 2kV continuous working voltage.
My requirements are;
Max Vpk continuous : 2kV
600Vrms CAT III Minimum (higher if possible)
1kVrms CAT II
I need to check the creepage and clearance in IEC61010 of the ACNV4506, at 13mm, I imagine this will be ok but I will make no assumptions.
So my real question is: What standards relate to the isolation transformers in switching PSUs that are used to power isolated measurement channels?
Do you know of any off the shelf solutions to meet the requirements of IEC61010 for a continuous working voltage of 2kVpk?
I have read in this article - (https://www.edn.com/Pdf/ViewPdf?contentItemId=4441292) It describes how IEC61010 gives guidance for safety tests of a product and how it details the creepage and clearances required from a product design perspective. But IEC61010 does not detail component specific requirements.
So as far as the isolation transformer for the power to the isolated board is concerned, other than the primary to secondary creepage and clearances from input to output leads, what standards are applicable to the overall isolation offered by a transformer?
I am not a transformer designer, is the only option to contact a transformer company to design a 2kV insulated transformer? Which standard should it meet?
Thanks
bode
I am designing a product which needs an isolated measurement input which will connect to mains (as well as various CATII circuits up to 1kVrms).
I am designing to IEC61010 and need to get communication across an isolation barrier as well as power across the isolation barrier to the analogue measurement circuits (as well as some digital circuitry).
IEC61010 describes the overall product testing requirements and creepage/clearance etc but it does not detail the requirements of the isolators themselves.
There is are a number of standards that apply to the opto isolators (I will use opto isolators to transfer data across the barrier), such as the IEC Standard is IEC60747-5-2.
The Broadcomm ACNV4506 opto is good for an operating voltage of 2262V, I need 2kV isolation so this is perfect.
However, I am not sure which standard to refer to for the isolated power.
I think I will need to design my own isolated PSU as i can’t find an off the shelf model offering a 2kV continuous working voltage.
My requirements are;
Max Vpk continuous : 2kV
600Vrms CAT III Minimum (higher if possible)
1kVrms CAT II
I need to check the creepage and clearance in IEC61010 of the ACNV4506, at 13mm, I imagine this will be ok but I will make no assumptions.
So my real question is: What standards relate to the isolation transformers in switching PSUs that are used to power isolated measurement channels?
Do you know of any off the shelf solutions to meet the requirements of IEC61010 for a continuous working voltage of 2kVpk?
I have read in this article - (https://www.edn.com/Pdf/ViewPdf?contentItemId=4441292) It describes how IEC61010 gives guidance for safety tests of a product and how it details the creepage and clearances required from a product design perspective. But IEC61010 does not detail component specific requirements.
So as far as the isolation transformer for the power to the isolated board is concerned, other than the primary to secondary creepage and clearances from input to output leads, what standards are applicable to the overall isolation offered by a transformer?
I am not a transformer designer, is the only option to contact a transformer company to design a 2kV insulated transformer? Which standard should it meet?
Thanks